has
08-25-2008, 11:49 AM
Just had a sales rep tell us that even though there is an external indicator on the pouches, we still need to be putting an indicator strip inside as well. True or false?

jrc
08-25-2008, 11:55 AM
We always do. AAMI ST79 Section 10, paragraph 5.2.2.2 says, "An internal CI shold be used within each package, tray, or rigid sterilization container system to be sterilized."

has
08-25-2008, 12:04 PM
Exactly what I needed to know. Thanks for the quick response.

anonymous
09-04-2008, 05:45 PM
Do you still need to put a indicator in the pouch if you are using instrument tip protectors that have an indicator on them?.

jrc
09-05-2008, 02:56 PM
If it is an approved indicator, I don't think it would matter what it looks like. I've never heard of tip protectors with indicators on them. Where do you get them?